Am J Clin Pathol
. 2020 Oct 5;aqaa181.
doi: 10.1093/ajcp/aqaa181. Online ahead of print.
A Comparison of Five SARS-CoV-2 Molecular Assays With Clinical Correlations
Gary W Procop 1 , Jay E Brock 1 , Edmunds Z Reineks 1 , Nabin K Shrestha 2 , Ryan Demkowicz 1 , Eleanor Cook 1 , Emad Ababneh 1 , Susan M Harrington 1
Affiliations
- PMID: 33015712
- DOI: 10.1093/ajcp/aqaa181
Abstract
Objectives: Comparative assessments of severe acute respiratory syndrome coronavirus 2 (SARS-CoV-2) molecular assays that have been operationalized through the US Food and Drug Administration's Emergency Use Authorization process are warranted to assess real-world performance. Characteristics such as sensitivity, specificity, and false-negative rate are important to inform clinical use.
Methods: We compared five SARS-CoV-2 assays using nasopharyngeal and nasal swab specimens submitted in transport media; we enriched this cohort for positive specimens, since we were particularly interested in the sensitivity and false-negative rate. Performance of each test was compared with a composite standard.
Results: The sensitivities and false-negative rates of the 239 specimens that met inclusion criteria were, respectively, as follows: Centers for Disease Control and Prevention 2019 nCoV Real-Time RT-PCR Diagnostic Panel, 100% and 0%; TIB MOLBIOL/Roche z 480 Assay, 96.5% and 3.5%; Xpert Xpress SARS-CoV-2 (Cepheid), 97.6% and 2.4%; Simplexa COVID-19 Direct Kit (DiaSorin), 88.1% and 11.9%; and ID Now COVID-19 (Abbott), 83.3% and 16.7%.
Conclusions: The assays that included a nucleic acid extraction followed by reverse transcription polymerase chain reaction were more sensitive than assays that lacked a full extraction. Most false negatives were seen in patients with low viral loads, as extrapolated from crossing threshold values.
Keywords: COVID-19; Coronavirus; Nucleic acid amplification tests; SARS-CoV-2.