• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Radiol Med. Proposal of a low-dose, long-pitch, dual-source chest CT protocol on third-generation dual-source CT using a tin filter for spectral shapi

tetano

Editor, Senior Moderator
Radiol Med. 2020 Apr 1. doi: 10.1007/s11547-020-01179-x. [Epub ahead of print]
Proposal of a low-dose, long-pitch, dual-source chest CT protocol on third-generation dual-source CT using a tin filter for spectral shaping at 100 kVp for CoronaVirus Disease 2019 (COVID-19) patients: a feasibility study.


Agostini A[SUP]1,[/SUP][SUP]2[/SUP], Floridi C[SUP]1,[/SUP][SUP]2[/SUP], Borgheresi A[SUP]3[/SUP], Badaloni M[SUP]2[/SUP], Esposto Pirani P[SUP]2[/SUP], Terilli F[SUP]2[/SUP], Ottaviani L[SUP]2[/SUP], Giovagnoni A[SUP]1,[/SUP][SUP]2[/SUP].

Author information




Abstract

AIM:

To subjectively and objectively evaluate the feasibility and diagnostic reliability of a low-dose, long-pitch dual-source chest CT protocol on third-generation dual-source CT (DSCT) with spectral shaping at 100Sn kVp for COVID-19 patients.
MATERIALS AND METHODS:

Patients with COVID-19 and positive swab-test undergoing to a chest CT on third-generation DSCT were included. The imaging protocol included a dual-energy acquisition (HD-DECT, 90/150Sn kVp) and fast, low-dose, long-pitch CT, dual-source scan at 100Sn kVp (LDCT). Subjective (Likert Scales) and objective (signal-to-noise and contrast-to-noise ratios, SNR and CNR) analyses were performed; radiation dose and acquisition times were recorded. Nonparametric tests were used.
RESULTS:

The median radiation dose was lower for LDCT than HD-DECT (Effective dose, ED: 0.28 mSv vs. 3.28 mSv, p = 0.016). LDCT had median acquisition time of 0.62 s (vs 2.02 s, p = 0.016). SNR and CNR were significantly different in several thoracic structures between HD-DECT and LDCT, with exception of lung parenchyma. Qualitative analysis demonstrated significant reduction in motion artifacts (p = 0.031) with comparable diagnostic reliability between HD-DECT and LDCT.
CONCLUSIONS:

Ultra-low-dose, dual-source, fast CT protocol provides highly diagnostic images for COVID-19 with potential for reduction in dose and motion artifacts.



KEYWORDS:

2019-nCOV; COVID-19; Chest radiology; Dual-source CT; Low-dose CT; Spectral shaping


PMID:32239472DOI:10.1007/s11547-020-01179-x
 