• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Pract Lab Med . Performance evaluation of influenza a rapid antigen test and PCR among nasopharyngeal and oropharyngeal samples

tetano

Editor, Senior Moderator
Pract Lab Med


. 2024 Jun 7:40:e00416.
doi: 10.1016/j.plabm.2024.e00416. eCollection 2024 May. Performance evaluation of influenza a rapid antigen test and PCR among nasopharyngeal and oropharyngeal samples

Xiaosong Su[SUP] 1 [/SUP], Jiaye Zhou[SUP] 1 2 [/SUP], Ling Liu[SUP] 1 [/SUP], Hongzhi Gao[SUP] 1 [/SUP], Yan Lin[SUP] 1 [/SUP], Zhile Wang[SUP] 1 [/SUP], Xin Zhang[SUP] 1 [/SUP], Baishen Pan[SUP] 2 [/SUP], Beili Wang[SUP] 2 [/SUP], Chunyan Zhang[SUP] 1 2 [/SUP], Wei Guo[SUP] 1 2 [/SUP]



Affiliations
Abstract

Objectives: Rapid antigen test (RAT) and polymerase chain reaction (PCR) using nasopharyngeal (NP) or oropharyngeal (OP) swab specimens are the two main testing techniques used for laboratory diagnosis of influenza in clinical practice. However, performance variations have been observed not only between techniques, but also between different specimens. This study evaluated the differences in performance between specimens and testing techniques to identify the best combination in clinical practice.
Methods: Both NP and OP samples from suspected influenza patients collected in the 2023/4-2023/5 Flu-season in Xiamen, China, were tested for RAT and quantitative PCR. The testing performance of the different specimens and testing techniques were recorded and evaluated.
Results: Compared to PCR, RAT showed 58.9 % and 10.3 % sensitivity for NP and OP swabs, respectively. The Limit of Detection (LoD) was 28.71 the Median Tissue Culture Infectious Dose (TCID[SUB]50[/SUB])/mL. Compared with PCR using NP swabs, PCR with OP swabs showed 89.5 % sensitivity and 95.4 % specificity.
Conclusions: There were no significant differences in performance between the specimens when PCR was used to test for influenza. However, a decrease in sensitivity was observed when the RAT was used, regardless of the specimen type. Therefore, to avoid false-negative results, PCR may be a better choice when OP swabs are used as specimens. In contrast, NP swabs should be the recommended specimens for RAT.

Keywords: Influenza; Nasopharyngeal; Oropharyngeal; Performance evaluation; Sample types.

 