• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Performance evaluation of a rapid molecular diagnostic, MultiCode based, sample-to-answer assay for the simultaneous detection of Influenza A, B and r

tetano

Editor, Senior Moderator
J Clin Virol. 2016 Nov 3;85:65-70. doi: 10.1016/j.jcv.2016.10.019. [Epub ahead of print]
[h=1]Performance evaluation of a rapid molecular diagnostic, MultiCode based, sample-to-answer assay for the simultaneous detection of Influenza A, B and respiratory syncytial viruses.[/h] Voermans JJ[SUP]1[/SUP], Seven-Deniz S[SUP]1[/SUP], Fraaij PL[SUP]2[/SUP], van der Eijk AA[SUP]1[/SUP], Koopmans MP[SUP]1[/SUP], Pas SD[SUP]3[/SUP].
[h=3]Author information[/h]

[h=3]Abstract[/h] [h=4]BACKGROUND:[/h] Clinical signs and symptoms of different airway pathogens are generally indistinguishable, making laboratory tests essential for clinical decisions regarding isolation and antiviral therapy. Immunochromatographic tests (ICT) and direct immunofluorescence assays (DFA) have lower sensitivities and specificities than molecular assays, but have the advantage of quick turnaround times and ease-of-use.
[h=4]OBJECTIVE:[/h] To evaluate the performance of a rapid molecular assay, ARIES FluA/B & RSV, using laboratory developed RT-PCR assays (LDA), ICT (BinaxNOW) and DFA.
[h=4]METHODS:[/h] Analytical and clinical performance were evaluated in a retrospective study arm (stored respiratory samples obtained between 2006-2015) and a prospective study arm (unselected fresh clinical samples obtained between December 2015 and March 2016 tested in parallel with LDAs).
[h=4]RESULTS:[/h] Genotype inclusivity and analytical specificity was 100%. However, ARIES was 0.5 log, 1-2logs and 2.5logs less sensitive for fluA, RSV and fluB respectively, compared to LDA. In total, 447 clinical samples were included, of which 15.4% tested positive for fluA, 9.2% for fluB and 26.0% for RSV, in both LDA and ARIES. ARIES clinical sensitivity compared to LDA was 98.6% (fluA), 93.3% (fluB) and 95.1% (RSV). Clinical specificity was 100% for all targets. ARIES detected 10.6% (4 fluA, 8 fluB, 11 RSV) and 26.9% (7 fluA, 3 fluB, 22 RSV) more samples compared to DFA and ICT, all confirmed by LDA.
[h=4]CONCLUSION:[/h] Although analytically ARIES is less sensitive than LDA, the clinical performance of the assay in our tertiary care setting was comparable, and significantly better than that of the established rapid assays.
Copyright ? 2016 Elsevier B.V. All rights reserved.


[h=4]KEYWORDS:[/h] Aries luminex; DFA; Direct immunofluorescence assay; ICT; Immunochromatography test; MultiCode; Point of impact; Rapid test; Real-time PCR

PMID: 27835760 DOI: 10.1016/j.jcv.2016.10.019