• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

J Clin Virol Validation and Verification of the Abbott RealTime SARS-CoV-2 Assay Analytical and Clinical Performance

tetano

Editor, Senior Moderator
J Clin Virol


. 2020 May 28;129:104474.
doi: 10.1016/j.jcv.2020.104474. Online ahead of print.
Validation and Verification of the Abbott RealTime SARS-CoV-2 Assay Analytical and Clinical Performance


Emily Degli-Angeli[SUP] 1 [/SUP], Joan Dragavon[SUP] 1 [/SUP], Meei-Li Huang[SUP] 2 [/SUP], Danijela Lucic[SUP] 3 [/SUP], Gavin Cloherty[SUP] 4 [/SUP], Keith R Jerome[SUP] 2 [/SUP], Alexander L Greninger[SUP] 2 [/SUP], Robert W Coombs[SUP] 5 [/SUP]



Affiliations

Abstract

Background: High-throughput assays for the SARS-CoV-2 virus are critical to increasing test capacity and slowing the spread of COVID-19. Abbott Molecular developed and received emergency use authorization (EUA) to deploy the new RealTime SARS-CoV-2 assay, run on the automated m2000sp/rt system.
Objective: To evaluate analytical and clinical performance of the RealTime SARS-CoV-2 assay compared to the SARS-CoV-2 CDC-based laboratory developed test (LDT) in clinical use by the University of Washington Clinical Virology Laboratory (UW Virology).
Methods: RealTime SARS-CoV-2 assay limit of detection (LOD) was evaluated by testing two dilution panels of 60 replicates each. Cross-reactivity was evaluated by testing 24 clinical samples positive for various non‒SARS-CoV-2 respiratory viruses. Clinical performance was evaluated using 30 positive and 30 negative SARS-CoV-2 clinical samples previously tested using the UW Virology SARS-CoV-2 LDT.
Results: Exceeding the 100 copies/mL LOD reported in the RealTime SARS-CoV-2 assay EUA product insert, 19 of 20 replicates were detected at 50 copies/mL and 16 of 20 replicates were detected at 25 copies/mL. All clinical samples positive for 24 non‒SARS-CoV-2 respiratory viruses were SARS-CoV-2 negative on the RealTime SARS-CoV-2 assay. The assay had high sensitivity (93%) and specificity (100%) for detecting SARS-CoV-2 in clinical samples. Two positive samples that tested negative with the RealTime SARS-CoV-2 assay had cycle numbers of 35.94 or greater and required dilution prior to testing. One of these samples was also inconclusive on the SARS-CoV-2 LDT.
Conclusion: The RealTime SARS-CoV-2 assay is acceptable for clinical use. With the high-throughput, fully automated m2000 system, this assay will accelerate the pace of SARS-CoV-2 testing.

Keywords: COVID-19; Coronavirus; EUA; High-throughput assay; PCR; SARS-CoV-2.
 