• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Infection . Evaluation of accuracy, exclusivity, limit-of-detection and ease-of-use of LumiraDx™: An antigen-detecting point-of-care device for SAR

tetano

Editor, Senior Moderator
Infection


. 2021 Aug 12.
doi: 10.1007/s15010-021-01681-y. Online ahead of print.
Evaluation of accuracy, exclusivity, limit-of-detection and ease-of-use of LumiraDx™: An antigen-detecting point-of-care device for SARS-CoV-2


Lisa J Krüger[SUP] 1 [/SUP], Julian A F Klein[SUP] 1 [/SUP], Frank Tobian[SUP] 1 [/SUP], Mary Gaeddert[SUP] 1 [/SUP], Federica Lainati[SUP] 1 [/SUP], Sarah Klemm[SUP] 2 [/SUP], Paul Schnitzler[SUP] 2 [/SUP], Ralf Bartenschlager[SUP] 3 4 [/SUP], Berati Cerikan[SUP] 3 [/SUP], Christopher J Neufeldt[SUP] 3 [/SUP], Olga Nikolai[SUP] 5 [/SUP], Andreas K Lindner[SUP] 5 [/SUP], Frank P Mockenhaupt[SUP] 5 [/SUP], Joachim Seybold[SUP] 6 [/SUP], Terry C Jones[SUP] 7 8 9 [/SUP], Victor M Corman[SUP] 7 8 [/SUP], Nira R Pollock[SUP] 10 [/SUP], Britta Knorr[SUP] 11 [/SUP], Andreas Welker[SUP] 11 [/SUP], Margaretha de Vos[SUP] 12 [/SUP], Jilian A Sacks[SUP] 12 [/SUP], Claudia M Denkinger[SUP] 13 14 [/SUP], ACE-IT Study Group



Collaborators, Affiliations

Abstract

Purpose: Rapid antigen-detecting tests (Ag-RDTs) for severe acute respiratory syndrome coronavirus 2 (SARS-CoV-2) can transform pandemic control. Thus far, sensitivity (≤ 85%) of lateral-flow assays has limited scale-up. Conceivably, microfluidic immunofluorescence Ag-RDTs could increase sensitivity for SARS-CoV-2 detection.
Methods: This multi-centre diagnostic accuracy study investigated performance of the microfluidic immunofluorescence LumiraDx™ assay, enrolling symptomatic and asymptomatic participants with suspected SARS-CoV-2 infection. Participants collected a supervised nasal mid-turbinate (NMT) self-swab for Ag-RDT testing, in addition to a professionally collected nasopharyngeal (NP) swab for routine testing with reverse transcriptase polymerase chain reaction (RT-PCR). Results were compared to calculate sensitivity and specificity. Sub-analyses investigated the results by viral load, symptom presence and duration. An analytical study assessed exclusivity and limit-of-detection (LOD). In addition, we evaluated ease-of-use.
Results: The study was conducted between November 2nd 2020 and 4th of December 2020. 761 participants were enrolled, with 486 participants reporting symptoms on testing day. 120 out of 146 RT-PCR positive cases were detected positive by LumiraDx™, resulting in a sensitivity of 82.2% (95% CI 75.2-87.5%). Specificity was 99.3% (CI 98.3-99.7%). Sensitivity was increased in individuals with viral load ≥ 7 log10 SARS-CoV2 RNA copies/ml (93.8%; CI 86.2-97.3%). Testing against common respiratory commensals and pathogens showed no cross-reactivity and LOD was estimated to be 2-56 PFU/mL. The ease-of-use-assessment was favourable for lower throughput settings.
Conclusion: The LumiraDx™ assay showed excellent analytical sensitivity, exclusivity and clinical specificity with good clinical sensitivity using supervised NMT self-sampling.
Trial registration number and registration date: DRKS00021220 and 01.04.2020.

Keywords: Antigen-detecting diagnostics; Covid-19; Diagnostic accuracy; Point-of-care; SARS-CoV-2.
 