Extremely high myoglobin plasma concentrations producing hook effect in a critically ill patient.
Kurt-Mangold M, Drees D, Krasowski MD.
Abstract
BACKGROUND:
A 21-year old female suffered a cardiac arrest after a one week history of viral illness later shown to be caused by influenza B. The patient required extended cardiopulmonary resuscitation and had further complications including compartment syndrome.
METHODS:
Plasma myoglobin concentration was measured using the Roche Diagnostics electrochemiluminescent myoglobin assay.
RESULTS:
The myoglobin concentration was 205,590μg/l in an undiluted specimen, consistent with severe rhabdomyolysis. Subsequent myoglobin concentrations measured two days later showed dramatic decreases to approximately 1000μg/l, raising suspicion of a hook effect. Dilution and re-analysis of the specimens revealed that the actual myoglobin concentrations were >395,000μg/l, with one specimen possessing an estimated myoglobin concentration of >600,000μg/l. Interestingly, three specimens from this patient did not show evidence of hook effect, with undiluted specimens producing myoglobin concentrations as high as 284,000μg/l. Retrospective analysis of myoglobin results over an 8-year period did not reveal other cases with suspicion of hook effect. The case patient had the highest myoglobin concentrations out of 7301 specimens.
CONCLUSIONS:
This case illustrates that while the Roche myoglobin assay has a very wide dynamic range, hook effect can occur with extremely high concentrations of plasma myoglobin.
Copyright ? 2012. Published by Elsevier B.V.
PMID:
22960204
[PubMed - as supplied by publisher]
http://www.ncbi.nlm.nih.gov/pubmed/22960204
Extremely high myoglobin plasma concentrations producing hook effect in a critically ill patient.
Kurt-Mangold M, Drees D, Krasowski MD.
Abstract
BACKGROUND:
A 21-year old female suffered a cardiac arrest after a one week history of viral illness later shown to be caused by influenza B. The patient required extended cardiopulmonary resuscitation and had further complications including compartment syndrome.
METHODS:
Plasma myoglobin concentration was measured using the Roche Diagnostics electrochemiluminescent myoglobin assay.
RESULTS:
The myoglobin concentration was 205,590μg/l in an undiluted specimen, consistent with severe rhabdomyolysis. Subsequent myoglobin concentrations measured two days later showed dramatic decreases to approximately 1000μg/l, raising suspicion of a hook effect. Dilution and re-analysis of the specimens revealed that the actual myoglobin concentrations were >395,000μg/l, with one specimen possessing an estimated myoglobin concentration of >600,000μg/l. Interestingly, three specimens from this patient did not show evidence of hook effect, with undiluted specimens producing myoglobin concentrations as high as 284,000μg/l. Retrospective analysis of myoglobin results over an 8-year period did not reveal other cases with suspicion of hook effect. The case patient had the highest myoglobin concentrations out of 7301 specimens.
CONCLUSIONS:
This case illustrates that while the Roche myoglobin assay has a very wide dynamic range, hook effect can occur with extremely high concentrations of plasma myoglobin.
Copyright ? 2012. Published by Elsevier B.V.
PMID:
22960204
[PubMed - as supplied by publisher]
http://www.ncbi.nlm.nih.gov/pubmed/22960204