• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Diagn Microbiol Infect Dis . Evaluation of the clinical performance of the JETest Influenza A/B Antigen Rapid Test Kit versus real-time reverse transc

tetano

Editor, Senior Moderator
Diagn Microbiol Infect Dis

. 2026 Aug 26;117(1):117629.
doi: 10.1016/j.diagmicrobio.2026.117629. Online ahead of print.

Evaluation of the clinical performance of the JETest Influenza A/B Antigen Rapid Test Kit versus real-time reverse transcription-PCR for influenza detection​


Yu-Tung Chen 1 , Ming-Hao Chen 2 , Kuan-Chen Chin 3 , Chiang-Ting Chien 4 , Thomas Yen-Shih Shen 5 , Chung-Chih Hung 6 , Ping-Hsun Yu 7

Affiliations


Abstract​


Background: The JETest Influenza A/B Antigen Rapid Test Kit is an in vitro diagnostic kit for detecting influenza A/B nucleoproteins in nasopharyngeal swab specimens. This study aims to verify the clinical performance of the kit compared with the real-time reverse transcription-PCR approach (the Cepheid Xpert® Xpress Flu/RSV assay kit).


Methods: Clinical specimens were collected using two nasopharyngeal swabs from each participant by healthcare professionals. One specimen was tested using a rapid influenza diagnostic test (RIDT), specifically the JETest Influenza A/B Antigen Rapid Test Kit; meanwhile, the result of the other specimen was confirmed by real-time reverse transcription-PCR (RT-PCR) to verify the true-positive or true-negative status of influenza A/B infection. Thereafter, the positive/negative percent agreements were calculated by comparing the test results of both methods.


Results: 118 patients were included in the study. For influenza A virus detection, the JETest demonstrated a positive percent agreement (PPA) of 85.5% (53/62) (95% CI: 74.2% - 93.1%) and a negative percent agreement (NPA) of 98.2% (55/56) (95% CI: 90.5% - 99.9%). For influenza B virus detection, the JETest demonstrated a PPA of 90.3% (28/31) (95% CI: 74.3% - 98.0%) and an NPA of 97.7% (85/87) (95% CI: 91.9% - 99.7%).


Conclusion: The clinical performance of the JETest influenza A/B Antigen Rapid Test Kit met the U.S. FDA clinical performance criteria for rapid influenza diagnostic tests and can serve as a reliable clinical aid in the diagnosis of influenza.

Keywords: Comparative Study; Lateral flow immunoassay; Rapid influenza diagnostic tests.
 