• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Comparison of three rapid influenza diagnostic tests with digital readout systems and one conventional rapid influenza diagnostic test

tetano

Editor, Senior Moderator
J Clin Lab Anal. 2017 Apr 13. doi: 10.1002/jcla.22234. [Epub ahead of print]
[h=1]Comparison of three rapid influenza diagnostic tests with digital readout systems and one conventional rapid influenza diagnostic test.[/h] Ryu SW[SUP]1[/SUP], Suh IB[SUP]1[/SUP], Ryu SM[SUP]2[/SUP], Shin KS[SUP]3[/SUP], Kim HS[SUP]4[/SUP], Kim J[SUP]5[/SUP], Uh Y[SUP]5[/SUP], Yoon KJ[SUP]5[/SUP], Lee JH[SUP]5[/SUP].
[h=3]Author information[/h]

[h=3]Abstract[/h] [h=4]BACKGROUND:[/h] Rapid influenza diagnostic tests (RIDTs) show variable sensitivities in clinical settings. We aimed to compare three digital RIDTs and one conventional RIDT.
[h=4]METHODS:[/h] We assessed 218 nasopharyngeal swabs from patients between neonates and 90 years old in 2016. Three digital RIDTs were BUDDI, Sofia Influenza A+B Fluorescence Immunoassay, Veritor System Flu A+B assay. One conventional test was the SD Bioline Influenza Ag A/B/A(H1N1/2009). All test results were compared with those from the Anyplex Flu A/B Typing Real-time Detection real-time PCR. The four RIDTs were tested with diluted solutions from the National Institute for Biological Standards and Control (NIBSC) to compare lower detection limit. Cross-reactivity of four RIDTs within other respiratory viruses was identified.
[h=4]RESULTS:[/h] For influenza A, BUDDI, Sofia, Veritor, and Bioline showed 87.7%, 94.5%, 87.7%, and 72.6% sensitivity, and 100%, 97.7%, 96.5%, and 100% specificity. For influenza B, BUDDI, Sofia, Veritor, and Bioline showed 81.7%, 91.7%, 81.7%, and 78.3% sensitivity, and 100%, 95.3%, 100%, and 100% specificity, respectively. Each RIDT could detect diluted NIBSC solution, according to the level of dilution and specific influenza subtypes. Cross-reactivity of four RIDTs with other respiratory viruses was not noted.
[h=4]CONCLUSIONS:[/h] Sofia showed the highest sensitivity for influenza A and B detection. BUDDI and Veritor showed higher detection sensitivity than a conventional RIDT for influenza A detection, but similar results for influenza B detection. Further study is needed to compare the test performance of RIDTs according to specific, prevalent influenza subtypes.
? 2017 Wiley Periodicals, Inc.


[h=4]KEYWORDS:[/h] digital readout system; influenza; rapid influenza diagnostic test; sensitivity; specificity

PMID: 28407318 DOI: 10.1002/jcla.22234
 