• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Clinical evaluation of multiplex RT-PCR assays for the detection of influenza A/B and respiratory syncytial virus using a high throughput system

tetano

Editor, Senior Moderator
J Virol Methods. 2019 Apr 1. pii: S0166-0934(18)30631-1. doi: 10.1016/j.jviromet.2019.03.015. [Epub ahead of print]
[h=1]Clinical evaluation of multiplex RT-PCR assays for the detection of influenza A/B and respiratory syncytial virus using a high throughput system.[/h] Eigner U[SUP]1[/SUP], Reucher S[SUP]2[/SUP], Hefner N[SUP]3[/SUP], Staffa-Peichl S[SUP]3[/SUP], Kolb M[SUP]3[/SUP], Betz U[SUP]3[/SUP], Holfelder M[SUP]3[/SUP], Spier G[SUP]4[/SUP], Pfefferle S[SUP]2[/SUP], L?tgehetmann M[SUP]2[/SUP].
[h=3]Author information[/h]

[h=3]Abstract[/h] [h=4]BACKGROUND:[/h] Lower respiratory tract infections are a major threat to public health systems worldwide, with RSV and influenza being the main agents causing hospitalization. In outbreak situations, high-volume respiratory testing is needed. In this study, we evaluated the analytical and clinical performance of a pre-designed primer/probe set for the simultaneous multiplex detection of both viruses on a high-throughput platform, the cobas 6800, using the "open channel" of the system for integration of lab-developed assays for the detection of influenza and RSV.
[h=4]RESULTS:[/h] Using the influenza/RSV qPCR Assay with swabs, LoD (95%) in TCID50/mL for influenza-A was 0.009, influenza-B 0.003, RSV-A 0.202, and RSV-B 0.009. Inter-run variability (3xLoD) was low (<1 Ct for all targets). Of 371 clinical respiratory specimens analyzed, results were concordant for 358 samples. The calculated sensitivity and specificity of the assay were 98.3% and 98.4% for Flu-A, 100% and 98.5% for Flu-B, and 98.6% and 99.7% for RSV. All quality assessment panel specimens (N = 63, including avian influenza strains) were correctly identified. None of the tested microorganisms showed cross-reactivity.
[h=4]CONCLUSION:[/h] Compared with CE-IVD assays, the assay evaluated here showed good analytical and clinical sensitivity and specificity with broad coverage of different virus strains. It offers high-throughput capacity with low hands-on time, facilitating the laboratory management of large respiratory outbreaks.
Copyright ? 2019. Published by Elsevier B.V.


[h=4]KEYWORDS:[/h] Influenza, Human; Multiplex Real-time PCR; Respiratory Syncytial Viruses

PMID: 30946852 DOI: 10.1016/j.jviromet.2019.03.015
 