• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Clin Chem. Clinical Performance of Two SARS-CoV-2 Serologic Assays

tetano

Editor, Senior Moderator
Clin Chem. 2020 May 13. pii: hvaa120. doi: 10.1093/clinchem/hvaa120. [Epub ahead of print]
Clinical Performance of Two SARS-CoV-2 Serologic Assays.


Tang MS[SUP]1[/SUP], Hock KG[SUP]1[/SUP], Logsdon NM[SUP]2[/SUP], Hayes JE[SUP]2[/SUP], Gronowski AM[SUP]1[/SUP], Anderson NW[SUP]1[/SUP], Farnsworth CW[SUP]1[/SUP].

Author information




Abstract

BACKGROUND:

The recent emergence of severe acute respiratory syndrome coronavirus 2 (SARS-CoV-2) has resulted in a rapid proliferation of serologic assays. However, little is known about their clinical performance. Here, we compared two commercial SARS-CoV-2 IgG assays.
METHODS:

103 specimens from 48 patients with PCR confirmed SARS-CoV-2 infections and 153 control specimens were analyzed using SARS-CoV-2 serologic assays by Abbott and EUROIMMUN (EI). Duration from symptom onset was determined by medical record review. Diagnostic sensitivity, specificity, and concordance were calculated.
RESULTS:

The Abbott SARS-CoV-2 assay had a diagnostic specificity of 99.4% (95% CI; 96.41-99.98%), and sensitivity of 0.0% (95% CI; 0.00-26.47%) at <3 days post symptom onset, 30.0% (95% CI; 11.89-54.28) at 3-7d, 47.8% (95% CI; 26.82-69.41) at 8-13d and 93.8% (95% CI; 82.80-98.69) at ≥14d. Diagnostic specificity on the EI assay was 94.8% (95% CI; 89.96-97.72) if borderline results were considered positive and 96.7% (95% CI; 92.54-98.93) if borderline results were considered negative. The diagnostic sensitivity was 0.0% (95% CI; 0.00-26.47%) at <3d, 25.0% (95% CI; 8.66-49.10) at 3-7d, 56.5% (95% CI; 34.49-76.81) at 3-7d and 85.4% (95% CI; 72.24-93.93) at ≥14d if borderline results were considered positive. The qualitative concordance between the assays was 0.83 (95% CI; 0.75-0.91).
CONCLUSIONS:

The Abbott SARS-CoV-2 assay had fewer false positive and false negative results than the EI assay. However, diagnostic sensitivity was poor in both assays during the first 14 days of symptoms.
? 2020 American Association for Clinical Chemistry.



KEYWORDS:

COVID-19; SARS-CoV-2; Serology


PMID:32402061DOI:10.1093/clinchem/hvaa120
 