doi: 10.1371/journal.pone.0329611. eCollection 2025. Diagnostic accuracy evaluation of a point-of-care antigen test for SARS-CoV-2 and influenza in UK primary care (RAPTOR-C19)
Thomas R Fanshawe[SUP] 1 [/SUP], Sharon Tonner[SUP] 1 [/SUP], Philip J Turner[SUP] 1 [/SUP], Margaret Głogowska[SUP] 1 [/SUP], Umasha Ukwatte[SUP] 1 [/SUP], Cecilia Okusi[SUP] 1 [/SUP], Jade Cogdale[SUP] 2 [/SUP], Maria Zambon[SUP] 3 4 [/SUP], Brian D Nicholson[SUP] 1 [/SUP], F D Richard Hobbs[SUP] 1 [/SUP], Gail N Hayward[SUP] 1 [/SUP]; RAPTOR-C19 Study Group
Affiliations
Objectives: To evaluate the diagnostic accuracy of the Roche SARS-CoV-2 & Flu A/B Rapid Antigen Test at the point of care.
Design: Prospective diagnostic accuracy study.
Setting: 17 primary care practices in England.
Participants: 500 individuals with symptoms consistent with possible SARS-CoV-2 or influenza infection identified upon presentation to primary care or via medical note review.
Primary and secondary outcome measures: Sensitivity, specificity and predictive values, compared to a laboratory reference standard of real-time reverse transcription PCR, using samples collected using a combined nasal and oropharyngeal swab.
Results: Of 481 participants with available index and reference test results, 5.6% (27/481) were reference standard-positive for SARS-CoV-2, 11.4% (55/481) for Influenza A and 1.9% (9/481) for Influenza B. The sensitivity of the antigen test to detect SARS-CoV-2 was 70.4% (19/27, 95% CI 49.6-86.2%) and specificity was 99.3% (451/454, 95%CI 98.1-99.9%). For Influenza A, sensitivity was 29.1% (16/55, 95% CI 17.6-42.9%) and specificity 98.6% (420/426, 97.0-99.5%), and for Influenza B, sensitivity was 22.2% (2/9, 2.8-60.0%) and specificity 98.1% (463/472, 96.4-99.1%).
Conclusions: In a primary care population of symptomatic individuals, the assay was highly specific and had moderate sensitivity to detect SARS-CoV-2, but did not detect the majority of influenza infections.
Registration: ISRCTN14226970.
. 2025 Aug 7;20(8):e0329611.
doi: 10.1371/journal.pone.0329611. eCollection 2025. Diagnostic accuracy evaluation of a point-of-care antigen test for SARS-CoV-2 and influenza in UK primary care (RAPTOR-C19)
Thomas R Fanshawe[SUP] 1 [/SUP], Sharon Tonner[SUP] 1 [/SUP], Philip J Turner[SUP] 1 [/SUP], Margaret Głogowska[SUP] 1 [/SUP], Umasha Ukwatte[SUP] 1 [/SUP], Cecilia Okusi[SUP] 1 [/SUP], Jade Cogdale[SUP] 2 [/SUP], Maria Zambon[SUP] 3 4 [/SUP], Brian D Nicholson[SUP] 1 [/SUP], F D Richard Hobbs[SUP] 1 [/SUP], Gail N Hayward[SUP] 1 [/SUP]; RAPTOR-C19 Study Group
Affiliations
- PMID: 40773492
- DOI: 10.1371/journal.pone.0329611
Objectives: To evaluate the diagnostic accuracy of the Roche SARS-CoV-2 & Flu A/B Rapid Antigen Test at the point of care.
Design: Prospective diagnostic accuracy study.
Setting: 17 primary care practices in England.
Participants: 500 individuals with symptoms consistent with possible SARS-CoV-2 or influenza infection identified upon presentation to primary care or via medical note review.
Primary and secondary outcome measures: Sensitivity, specificity and predictive values, compared to a laboratory reference standard of real-time reverse transcription PCR, using samples collected using a combined nasal and oropharyngeal swab.
Results: Of 481 participants with available index and reference test results, 5.6% (27/481) were reference standard-positive for SARS-CoV-2, 11.4% (55/481) for Influenza A and 1.9% (9/481) for Influenza B. The sensitivity of the antigen test to detect SARS-CoV-2 was 70.4% (19/27, 95% CI 49.6-86.2%) and specificity was 99.3% (451/454, 95%CI 98.1-99.9%). For Influenza A, sensitivity was 29.1% (16/55, 95% CI 17.6-42.9%) and specificity 98.6% (420/426, 97.0-99.5%), and for Influenza B, sensitivity was 22.2% (2/9, 2.8-60.0%) and specificity 98.1% (463/472, 96.4-99.1%).
Conclusions: In a primary care population of symptomatic individuals, the assay was highly specific and had moderate sensitivity to detect SARS-CoV-2, but did not detect the majority of influenza infections.
Registration: ISRCTN14226970.