• FluTrackers.com Inc. does not provide medical advice. Information on this web site is collected from various internet resources, and the FluTrackers board of directors makes no warranty to the safety, efficacy, correctness or completeness of the information posted on this site by any author or poster. The information collated here is for instructional and/or discussion purposes only and is NOT intended to diagnose or treat any disease, illness, or other medical condition. Every individual reader or poster should seek advice from their personal physician/healthcare practitioner before considering or using any interventions that are discussed on this website. By continuing to access this website you agree to consult your personal physican before using any interventions posted on this website, and you agree to hold harmless FluTrackers.com Inc., the board of directors, the members, and all authors and posters for any effects from use of any medication, supplement, vitamin or other substance, device, intervention, etc. mentioned in posts on this website, or other internet venues referenced in posts on this website.
  • We are not asking for any donations. Do not donate to any entity who says they are raising funds for us.

Lancet Microbe . Assessing a novel, lab-free, point-of-care test for SARS-CoV-2 (CovidNudge): a diagnostic accuracy study

tetano

Editor, Senior Moderator
Lancet Microbe


. 2020 Sep 17.
doi: 10.1016/S2666-5247(20)30121-X. Online ahead of print.
Assessing a novel, lab-free, point-of-care test for SARS-CoV-2 (CovidNudge): a diagnostic accuracy study


Malick M Gibani[SUP] 1 2 [/SUP], Christofer Toumazou[SUP] 3 4 [/SUP], Mohammadreza Sohbati[SUP] 3 [/SUP], Rashmita Sahoo[SUP] 3 [/SUP], Maria Karvela[SUP] 3 [/SUP], Tsz-Kin Hon[SUP] 3 [/SUP], Sara De Mateo[SUP] 3 [/SUP], Alison Burdett[SUP] 4 [/SUP], K Y Felice Leung[SUP] 3 [/SUP], Jake Barnett[SUP] 3 [/SUP], Arman Orbeladze[SUP] 3 [/SUP], Song Luan[SUP] 3 [/SUP], Stavros Pournias[SUP] 3 [/SUP], Jiayang Sun[SUP] 3 [/SUP], Barney Flower[SUP] 1 2 [/SUP], Judith Bedzo-Nutakor[SUP] 3 [/SUP], Maisarah Amran[SUP] 2 [/SUP], Rachael Quinlan[SUP] 1 [/SUP], Keira Skolimowska[SUP] 1 2 [/SUP], Carolina Herrera[SUP] 1 [/SUP], Aileen Rowan[SUP] 1 [/SUP], Anjna Badhan[SUP] 1 [/SUP], Robert Klaber[SUP] 2 [/SUP], Gary Davies[SUP] 5 [/SUP], David Muir[SUP] 2 [/SUP], Paul Randell[SUP] 2 [/SUP], Derrick Crook[SUP] 6 [/SUP], Graham P Taylor[SUP] 1 [/SUP], Wendy Barclay[SUP] 1 [/SUP], Nabeela Mughal[SUP] 1 2 5 [/SUP], Luke S P Moore[SUP] 1 2 5 [/SUP], Katie Jeffery[SUP] 6 7 [/SUP], Graham S Cooke[SUP] 1 2 [/SUP]



Affiliations

Abstract

Background: Access to rapid diagnosis is key to the control and management of severe acute respiratory syndrome coronavirus 2 (SARS-CoV-2). Laboratory RT-PCR testing is the current standard of care but usually requires a centralised laboratory and significant infrastructure. We describe our diagnostic accuracy assessment of a novel, rapid point-of-care real time RT-PCR CovidNudge test, which requires no laboratory handling or sample pre-processing.
Methods: Between April and May, 2020, we obtained two nasopharyngeal swab samples from individuals in three hospitals in London and Oxford (UK). Samples were collected from three groups: self-referred health-care workers with suspected COVID-19; patients attending emergency departments with suspected COVID-19; and hospital inpatient admissions with or without suspected COVID-19. For the CovidNudge test, nasopharyngeal swabs were inserted directly into a cartridge which contains all reagents and components required for RT-PCR reactions, including multiple technical replicates of seven SARS-CoV-2 gene targets (rdrp1, rdrp2, e-gene, n-gene, n1, n2 and n3) and human ribonuclease P (RNaseP) as sample adequacy control. Swab samples were tested in parallel using the CovidNudge platform, and with standard laboratory RT-PCR using swabs in viral transport medium for processing in a central laboratory. The primary analysis was to compare the sensitivity and specificity of the point-of-care CovidNudge test with laboratory-based testing.
Findings: We obtained 386 paired samples: 280 (73%) from self-referred health-care workers, 15 (4%) from patients in the emergency department, and 91 (23%) hospital inpatient admissions. Of the 386 paired samples, 67 tested positive on the CovidNudge point-of-care platform and 71 with standard laboratory RT-PCR. The overall sensitivity of the point-of-care test compared with laboratory-based testing was 94% (95% CI 86-98) with an overall specificity of 100% (99-100). The sensitivity of the test varied by group (self-referred healthcare workers 93% [95% CI 84-98]; patients in the emergency department 100% [48-100]; and hospital inpatient admissions 100% [29-100]). Specificity was consistent between groups (self-referred health-care workers 100% [95% CI 98-100%]; patients in the emergency department 100% [69-100]; and hospital inpatient admissions 100% [96-100]). Point of care testing performance was similar during a period of high background prevalence of laboratory positive tests (25% [95% 20-31] in April, 2020) and low prevalence (3% [95% 1-9] in inpatient screening). Amplification of viral nucleocapsid (n1, n2, and n3) and envelope protein gene (e-gene) were most sensitive for detection of spiked SARS-CoV-2 RNA.
Interpretation: The CovidNudge platform was a sensitive, specific, and rapid point of care test for the presence of SARS-CoV-2 without laboratory handling or sample pre-processing. The device, which has been implemented in UK hospitals since May, 2020, could enable rapid decisions for clinical care and testing programmes.
Funding: National Institute of Health Research (NIHR) Imperial Biomedical Research Centre, NIHR Health Protection Research Unit in Healthcare Associated Infections and Antimicrobial Resistance at Oxford University in partnership with Public Health England, NIHR Biomedical Research Centre Oxford, and DnaNudge.
 