Diagnosis (Berl)
. 2023 Jan 20.
doi: 10.1515/dx-2022-0138. Online ahead of print.
Real-world assessment of the clinical performance of COVID-VIRO ALL IN rapid SARS-CoV-2 antigen test
Laura Pighi[SUP] 1 2 [/SUP], Brandon M Henry[SUP] 3 [/SUP], Simone De Nitto[SUP] 1 2 [/SUP], Gianluca Gianfilippi[SUP] 4 [/SUP], Gian Luca Salvagno[SUP] 1 2 [/SUP], Giuseppe Lippi[SUP] 1 [/SUP]
Affiliations
Abstract
Objectives: Since the external validation of severe acute respiratory syndrome coronavirus 2 antigen rapid diagnostic tests (SARS-CoV-2 RDT-Ags) is a necessary requisite before they can be introduced into routine clinical practice, this study reports the results of a real-world assessment of the clinical performance of the new COVID-VIRO ALL IN device.
Methods: The study population consisted in 165 outpatients (median age: 43 years, range: 14-68 years; 66.1% females) who had paired nasal and nasopharyngeal samples collected upon hospital presentation. The samples were concomitantly tested with the AAZ-LMB COVID-VIRO ALL IN SARS-CoV-2 RDT-Ag and with Cepheid Xpert Xpress SARS-CoV-2 real-time reverse transcription polymerase chain reaction (RT-PCR).
Results: The number of subjects with positive RT-PCR results (i.e., mean Ct value <45) was 116 (70.3%), 109 (66.1%) and 86 (52.1%) with mean Ct values <37 and <30, respectively. In all RT-PCR positive samples, COVID-VIRO ALL IN displayed 78.8% agreement, 0.698 sensitivity, 1.000 specificity, 0.583 negative predictive value (NPV) and 1.000 positive predictive value (PPV) compared to RT-PCR. The median Ct value of samples testing positive with COVID-VIRO ALL IN was significantly lower than those testing negative (22.8 vs. 32.2; p<0.001). In samples with high viral load (i.e., Ct value <30), COVID-VIRO ALL IN displayed 92.1% agreement, 0.895 sensitivity, 0.949 specificity, 0.983 NPV and 0.951 PPV compared to RT-PCR.
Conclusions: Although the diagnostic performance of COVID-VIRO ALL IN do not exactly match those of the manufacturer, its high NPV in high viral load samples would enable fast-track and rapid identification of highly contagious subjects.
Keywords: COVID-19; SARS-CoV-2; antigen testing; diagnosis; infection.
. 2023 Jan 20.
doi: 10.1515/dx-2022-0138. Online ahead of print.
Real-world assessment of the clinical performance of COVID-VIRO ALL IN rapid SARS-CoV-2 antigen test
Laura Pighi[SUP] 1 2 [/SUP], Brandon M Henry[SUP] 3 [/SUP], Simone De Nitto[SUP] 1 2 [/SUP], Gianluca Gianfilippi[SUP] 4 [/SUP], Gian Luca Salvagno[SUP] 1 2 [/SUP], Giuseppe Lippi[SUP] 1 [/SUP]
Affiliations
- PMID: 36656638
- DOI: 10.1515/dx-2022-0138
Abstract
Objectives: Since the external validation of severe acute respiratory syndrome coronavirus 2 antigen rapid diagnostic tests (SARS-CoV-2 RDT-Ags) is a necessary requisite before they can be introduced into routine clinical practice, this study reports the results of a real-world assessment of the clinical performance of the new COVID-VIRO ALL IN device.
Methods: The study population consisted in 165 outpatients (median age: 43 years, range: 14-68 years; 66.1% females) who had paired nasal and nasopharyngeal samples collected upon hospital presentation. The samples were concomitantly tested with the AAZ-LMB COVID-VIRO ALL IN SARS-CoV-2 RDT-Ag and with Cepheid Xpert Xpress SARS-CoV-2 real-time reverse transcription polymerase chain reaction (RT-PCR).
Results: The number of subjects with positive RT-PCR results (i.e., mean Ct value <45) was 116 (70.3%), 109 (66.1%) and 86 (52.1%) with mean Ct values <37 and <30, respectively. In all RT-PCR positive samples, COVID-VIRO ALL IN displayed 78.8% agreement, 0.698 sensitivity, 1.000 specificity, 0.583 negative predictive value (NPV) and 1.000 positive predictive value (PPV) compared to RT-PCR. The median Ct value of samples testing positive with COVID-VIRO ALL IN was significantly lower than those testing negative (22.8 vs. 32.2; p<0.001). In samples with high viral load (i.e., Ct value <30), COVID-VIRO ALL IN displayed 92.1% agreement, 0.895 sensitivity, 0.949 specificity, 0.983 NPV and 0.951 PPV compared to RT-PCR.
Conclusions: Although the diagnostic performance of COVID-VIRO ALL IN do not exactly match those of the manufacturer, its high NPV in high viral load samples would enable fast-track and rapid identification of highly contagious subjects.
Keywords: COVID-19; SARS-CoV-2; antigen testing; diagnosis; infection.