The U.S. Food and Drug Administration (FDA) is alerting clinical laboratory staff, point-of-care (POC) facility staff, and health care providers that false positive results can occur with the Roche Molecular Systems, Inc. (Roche) cobas SARS-CoV-2 & Influenza A/B Nucleic Acid Test for use on the cobas Liat System. The false positive results may be related to two issues observed by Roche.
The U.S. Food and Drug Administration (FDA) is alerting clinical laboratory staff, point-of-care (POC) facility staff, and health care providers that false positive results can occur with the Roche Molecular Systems, Inc. (Roche) cobas SARS-CoV-2 & Influenza A/B Nucleic Acid Test for use on the cobas Liat System. The false positive results may be related to two issues observed by Roche.
- Roche identified that the assay tubes may sporadically leak, causing an obstructed optical path in the Liat analyzer, producing abnormal PCR growth curves. This could lead to invalid or erroneous positive results, particularly for the Flu B test. If a tube leak occurs, later testing runs may have an increased likelihood of false positive Flu B results.
- Roche determined that abnormal PCR cycling in the reaction tubes may also produce abnormal PCR growth curves, leading to erroneous results. The issue is sporadic and may be caused by multiple factors happening at the same time, such as hardware positioning, volume movement, and curve interpretation. This issue may cause false positive results for multiple analytes (Influenza A, Influenza B and/or SARS-CoV-2) in a single testing run.